Default Fluke 333-337 Digital Clamp Meter Recall

See:

Digital Clamp Meters Recalled by Fluke Due to Shock Hazard


If you have one of these, you know what to do.

ST
Reply With Quote
Old May 20th, 2009, 10:01 PM
derekleffew's Avatar
Senior Team
 Premium Member 
 
Join Date: Aug 2007
Location: Las Vegas, NV, USA
Posts: 3,535
Thanks: 98
Thanked 261 Times in 227 Posts
Blog Entries: 1
Default Re: Fluke 333-337 Digital Clamp Meter Recall

Not as bad as it sounds, really. From Fluke test tools, answering your electronic, electrical, predictive maintenance and biomedical needs. :
Quote:
Through quality assurance tests, we have discovered a potential short circuit connection on the circuit board. This may lead to inaccurate voltage readings including a low or no voltage reading on a circuit energized with a hazardous voltage. This finding could create a hazardous situation if the user were to contact live voltage based on an erroneous instrument reading.
So what's the worst that could happen if a meter tells you a circuit is dead when it is, in fact, energized? Electrocution?
__________________
"It's a shame there couldn't have been more LED fixtures at this year's LDI."
.
Reply With Quote
Old May 20th, 2009, 11:57 PM

 
Join Date: Apr 2008
Location: Minneapolis, MN
Posts: 198
Thanks: 0
Thanked 4 Times in 4 Posts
Default Re: Fluke 333-337 Digital Clamp Meter Recall

Good reason when checking for dead to go with the live-dead-live test...verify your meter / tick tracer's operation on a known live circuit, THEN check the circuit in question, then check the known live circuit again...this is how I learned at the nuclear power plant I used to work at, I assume it's a recommendation in NFPA 70E but that's only a conjecture.
__________________
Many techs only go so far as to prevent problems that can be blamed on them...I feel it is only right to prevent ANY problem that I can that has the potential to take down a show.
Reply With Quote
Old May 21st, 2009, 11:46 AM

 
Join Date: Feb 2009
Location: SW Ontario, Canada
Posts: 189
Thanks: 0
Thanked 4 Times in 4 Posts
Default Re: Fluke 333-337 Digital Clamp Meter Recall

I was taught in trade-school to do the same (always verify your meter's reading on a circuit of known voltage!). It can save your life.
Also NEVER trust those inductive pen-testers. They lie ALL the time.

I've had them tell me dead circuits were live, and live circuits were dead. Good thing I usually have a Digital-Multi-Meter around.

On a side-note:

Multimeters (usually) never come with fused leads, I defiantly recommend anyone using one should get and use fused leads. Many times I've seen multimeters blow sky high without them!!!!

And as-per arc-flash training, many electricians say that they "never work live". If you never work live, you clearly never use your meter.
Reply With Quote
Old May 28th, 2009, 03:25 PM
MNicolai's Avatar
 Premium Member 

 
Join Date: Mar 2008
Location: Oconomowoc, WI
Posts: 487
Thanks: 26
Thanked 23 Times in 22 Posts
Default Re: Fluke 333-337 Digital Clamp Meter Recall

From a Fluke employee @ Test and Measurement Tool Users Community - Fused Test Leads regarding fused test leads in exclusive relation to the use of Fluke meters, and applies not necessisarily to any other brand or product.

Quote:
Here is our opinion on fused test leads:

It is Fluke's opinion that in applications using Fluke name brand instruments, fused test probes provide little if any added protection. Fluke's primary concern is that fused test probes may result in serious shock hazard do to a false voltage reading. A more effective solution is a meter properly designed to handle overload conditions. For the Amp and milliamp functions, Fluke meters incorporate a specifically designed and tested fuse that is effective to 1000 Volts as the full breaking capacity of the system which limits the energy in a manner that is as effective as a fuse located within a test probe.



It is additionally Fluke's position that a fused test probe is potentially more hazardous when used with any DMM for the following reason. It is estimated that over 90% of most electrician measurements use the AC VOLTS functions of a meter. Current measures are generally less than 10%, more probably in the few percentage points. If the fuse in the test probe is damaged and not noticed or properly corrected, then false voltage measurements will occur which potentially create a shock hazard for the user.



For Fluke meters, a damaged fuse within the amps function of the meter does not affect voltage measurements, so the potential of shock hazard due to false readings is eliminated. Note that for the other meter functions such as ohms, Fluke meters are designed to prevent hazard for overload conditions equivalent the maximum AC volts input as stated on the meter. These functions limit the energy by impedance. This opinion on fused test leads is limited to, an only applies to their use when used with Fluke brand test instruments

Duane
__________________
Mike Nicolai
Oconomowoc, WI, USA
mike skims his id on twitter
mike talks about things that matter to him on tumblr
Reply With Quote
Reply

Tags
337, clamp, digital, fluke, meter, recall

Thread Tools Search this Thread
Search this Thread:

Advanced Search
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is On
Trackbacks are On
Pingbacks are On
Refbacks are On


Similar Threads
Thread Thread Starter Forum Replies Last Post
Lighting Fixture Maintenance derekleffew Collaborative Articles 0 April 25th, 2009 12:29 AM
Clamp Meter derekleffew Glossary A-E 0 June 9th, 2008 04:04 AM
Clamp Meters Diarmuid Lighting 66 February 16th, 2008 01:18 AM
standard instrument maintenance moojoe General Advice 13 May 11th, 2005 11:31 AM


All times are UTC -4. The time now is 11:23 AM.


Powered by vBulletin® Version 3.8.4
Copyright ©2000 - 2009, Jelsoft Enterprises Ltd.
Search Engine Friendly URLs by vBSEO 3.3.1 
Advertisement System V2.6 By   Branden

1 2 3 4 5 6 7 8 9 10 11 12 13 14 15 16 17 18 19 20 21 22 23 24 25 26 27 28 29 30 31 32 33 34 35 36 37 38 39 40 41 42 43 44 45 46 47 48 49 50 51 52 53 54 55 56 57 58 59 60 61 62 63 64 65 66 67 68 69 70 71 72 73 74 75 76 77 78 79 80